Leica Cm1850 Cryostat - Answered Questions & Fixed issues

0helpful
1answer

Cryostat no longer cools

Try to check the relays. There is what we call a protective thermostat and thermistor for the compressor. Maybe this is also busted. A starting thermistor usually gives way in todays compressors. You can see this in the connector of the compressor.
7/7/2014 8:00:29 PM • Leica Cm1850... • Answered on Jul 07, 2014 • 377 views
0helpful
1answer

How to get my microscope to turn on.

Richard: Double-check the electrical source (plug) you have it plugged into with another device. Just because the bulb "looks good" does NOT mean it IS good, also. If the electrical source ("the plug") is "hot", then the likely culprit is the bulb. Next would be the electrical switch - "fiddle" with it a bit. (work it like a toilet that is problematic: jiggle the handle. wiggle things a bit. yell at it. whack it a few times) If that doesn't work, check your warranty to see if a repair is covered. (don't "whack it" so hard the warranty is void!)
8/26/2014 8:34:26 PM • Leica Science &... • Answered on Aug 26, 2014 • 268 views
0helpful
1answer

We have got a leica

This issue could potentially be solved by a flatfield software correction of the camera that you are using. I would recommend visiting leica-microsystems.com and filling out a Contact Me request. Your local Leica specialist will then be in touch to help.
1/4/2011 2:46:15 PM • Leica Science &... • Answered on Jan 04, 2011 • 295 views
0helpful
1answer

Need for leica dm5000b user manual

4/12/2010 11:15:21 PM • Leica Science &... • Answered on Apr 12, 2010 • 456 views
0helpful
1answer

Do we need to periodically calibrate the

Depends on the microscope. Generally, with most simple microscopes that are used often, calibration (aligning of the prisms, objectives and eyepieces) shouldn't be required more than once every 5 years or so.
4/22/2014 4:16:43 PM • Leica Science &... • Answered on Apr 22, 2014 • 368 views
0helpful
1answer

Need manual for Leica MZ16F

Tech specs and exploded diagram http://www.clemex.com/pdf/brochures/Leica_MZ_16_Tech_en.pdf
Uusers manual http://biotech.wku.edu/online_manual/Leica%20MZ16%20Fluorescence%20Stereomicroscope.pdf
Enjoy
10/12/2009 3:38:03 AM • Leica Science &... • Answered on Oct 12, 2009 • 891 views
0helpful
1answer

Compare Apex Practitioner Microscope v C & A Scientific MFL-06 My First Lab Duo-Scope?

You are probably qualified by now but you question was resurrected by and AI Bot. https://www.google.com/search?q=Apex+Practitioner+Microscope+vs+C+%26+A+Scientific+MFL-06 ..
12/30/2023 9:34:23 PM • Science &... • Answered on Dec 30, 2023 • 119 views
0helpful
1answer

Little Sister 3 Autoclave is displaying FILL but does not actually fill and goes no further. What can I do?

Technical Manual and other suggestions. Professional machines really need professional service as the repair is very specialized, and the dangers of using incorrectly sterilized tools injurious to health.

T echnical Manual

Frank's Hospital Workshop http://www.frankshospitalworkshop.com > Esch... The SES Little Sister 3 is a portable electrically operated steam autoclave suitable for the sterilization of unwrapped metal instruments and utensils. 4. The ... 47 pages

Watch out for These 7 Common Autoclave Problems

Consolidated Sterilizer Systems https://consteril.com > common-autoclave-problems 1. The sterilizer won't reach the temperature set-point. · The steam generator is need of repair. · There is a clog in the chamber drain (i.e. check the chamber ... Missing: Sister ‎' Show results with: Sister
11/19/2023 6:33:32 PM • Science &... • Answered on Nov 19, 2023 • 55 views
0helpful
2answers
0helpful
1answer

How will you track moving image in a compound binocular microscope?

10/20/2023 2:12:33 PM • Science &... • Answered on Oct 20, 2023 • 845 views
0helpful
1answer

Eppendorf 5402 centrifuge front screen not working

you are right, there is no manual, except for the 5420. Try here, they do Sales and Service as well https://www.gmi-inc.com/product/eppendorf-5402-refrigerated-centrifuge/ and you can still buy them https://angelusmedical.com/products/eppendorf-5402-refrigerated-micro-centrifuge-parts-only Did you ask them direct for support? Did you open it up and check Ribbon Cable connections?

Centrifuge and Rotor Service

Eppendorf https://www.eppendorf.com > ca-en > service-support epServices offers rotor assurance plans. Read more about the professional inspection services and evaluation of centrifuge rotors and buckets.

Service & Support

Eppendorf https://www.eppendorf.com > ca-en > service-support The services range from technical support and troubleshooting to delivery of replacement parts on short notice and customer-specific maintenance programs. ULT ...

Service & Support

Eppendorf https://www.eppendorf.com > your-centrifuge-solutions We provide efficient, reliable services for your centrifuge. Our ... Includes a full annual preventive maintenance service program for assured instrument ...
10/11/2023 7:38:15 PM • Science &... • Answered on Oct 11, 2023 • 56 views
0helpful
1answer

What is the E009 problem and how to repair please.

The Autoclave door is not opening at the end of the cycle.

Midmark M11 "Old Style" Error Code E009

Sterilizer Autoclave Solutions https://statimusa.com > midmark-m11-old-style-error-... Error E009: Door doesn't fully open at the end of a cycle. This error indicates that there is pressure buildup during the dry cycle. Probable causes of Midmark ...

Error Code E009 :

AllClaveParts https://www.allclaveparts.com > error-code-e009 You will need to know whether or not the Door is Opening Automatically at the End of the Cycle when you get the Error Code E009 before proceeding.
4/1/2023 3:34:52 AM • Science &... • Answered on Apr 01, 2023 • 142 views
0helpful
1answer

AA-Spectrometer Unicam 939 Error

The error you are experiencing on your Unicam 939 AA-Spectrometer with the software reporting error 16 and 2, along with lamps 1,2,3,4, and 10 lighting up, could indicate a few different issues. Here are some steps you can take to try to fix the problem:
  1. Restart the spectrometer: Turn off the spectrometer and unplug it from the power source. Wait a few minutes, then plug it back in and turn it on again. Sometimes, a simple restart can fix issues with the instrument.
  2. Check the calibration: It is possible that the instrument needs to be calibrated. Follow the manufacturer's instructions for calibrating the spectrometer.
  3. Check the lamps: If multiple lamps are lighting up, it could be due to a problem with the lamp assembly. Check that the lamps are properly installed and functioning. Try replacing any lamps that appear damaged or worn.
  4. Check the power supply: Make sure the power supply is functioning properly by testing the voltage output with a multimeter. If the voltage output is incorrect, try replacing the power supply.
  5. Check the communication between the software and the spectrometer: Ensure that the software is properly communicating with the spectrometer. Check that the cables are properly connected and that the software settings are correct.
If these steps do not resolve the issue, it may be necessary to contact the manufacturer for additional support or to seek the assistance of a professional repair technician.
3/24/2023 6:39:35 AM • Science &... • Answered on Mar 24, 2023 • 67 views
0helpful
1answer

Eppendorf centrifuge model 5702: lid not does latch/lock after emergency release.

If the lid of your Eppendorf centrifuge model 5702 is not latching or locking closed after the emergency release, it is possible that the locking mechanism is damaged or misaligned. Here are some steps you can try to troubleshoot the issue:
  1. Check the alignment of the lid: Make sure that the lid is properly aligned with the base of the centrifuge. If it is not, the locking mechanism may not engage correctly. Adjust the lid as needed and try to latch it closed.
  2. Inspect the locking mechanism: Check the locking mechanism on both the lid and the base of the centrifuge for any signs of damage or wear. If you see any damage, you may need to replace the locking mechanism.
  3. Check the safety interlock switch: The safety interlock switch is a small button located on the base of the centrifuge that is depressed when the lid is closed. If this switch is not working correctly, it can prevent the lid from latching closed. Check that the switch is functioning properly by pressing it with a small tool while attempting to latch the lid.
  4. Contact Eppendorf customer support: If none of the above steps work, you may need to contact Eppendorf customer support for further assistance. They can help you troubleshoot the issue and provide advice on how to proceed.
3/15/2023 4:55:19 AM • Science &... • Answered on Mar 15, 2023 • 194 views
0helpful
1answer

Enigma 8 autoclave error code E13

The E13 error code on an Enigma 8 autoclave typically indicates a problem with the vacuum system. This error code is indicating that the vacuum is not reaching the proper level during the sterilization cycle. Here are a few things you can check to fix the E13 error code:
  1. Check the vacuum pump: Make sure that the vacuum pump is functioning properly and that it is not blocked or clogged.
  2. Check the air filter: Make sure that the air filter is clean and not clogged. A clogged air filter can prevent the vacuum pump from functioning properly.
  3. Check the vacuum hose: Make sure that the vacuum hose is not damaged or disconnected.
  4. Check the door gasket: Make sure that the door gasket is not damaged or leaking. A damaged or leaking door gasket can prevent the vacuum from reaching the proper level.
  5. Check the control board: Make sure that the control board is not damaged or malfunctioning.
  6. Check the water level: Make sure that the water level is at the correct level in the autoclave chamber.
1/14/2023 10:21:30 PM • Science &... • Answered on Jan 14, 2023 • 1,449 views
0helpful
1answer

Zeus's Acuitus 5015 only reads left eye.

Contact Carl Zeiss direct. That is well beyond the scope of Fixya, sorry. https://www.google.com/search?q=Zeus%27s+Acuitus+5015 ..
1/5/2023 6:23:50 PM • Science &... • Answered on Jan 05, 2023 • 61 views
0helpful
1answer

Ernst Leitz Wetzlar Laborlux K Microscope. The condenser on my Laborlux K is locked in the top position. How do I unlock it?

A bit specialized for this site but page 7 onwards of the manual covers all the condenser unit types.http://earth2geologists.net/Microscopes/documents/Leitz_LaborluxS_instructions.pdf..
12/22/2022 6:28:15 PM • Science &... • Answered on Dec 22, 2022 • 15 views
Not finding what you are looking for?
Leica Cm1850 Cryostat Logo

3 questions posted

Ask a Question

Usually answered in minutes!

Top Leica Science & Laboratory Experts

ADMIN Andrew
ADMIN Andrew

Level 3 Expert

66931 Answers

Collins Mccoy
Collins Mccoy

Level 2 Expert

58 Answers

Insta lab
Insta lab

Level 1 Expert

5 Answers

Are you a Leica Science and Laboratory Expert? Answer questions, earn points and help others

Answer questions

Manuals & User Guides

Loading...