SOURCE: fujitsu 4860c scanner error
There are 2 sensors on the background unit(Front) SF1.5 and SF 2
Both sensors look like black boxes. SF2 is nearest the reading glass and SF1.5 is below it. Make sure both are clean. There are also tests you can run in test mode to recalibrate the sensors, if that doesn't work you have to replace it.
Posted on Aug 26, 2008
  • 32 Answers
SOURCE: Fujitsu fi 4750c scanner error
Try a flatbed scan scan with the cover open.  If you see the lamp come on, it's likely a cable or CCD problem, though it can be caused by a really dirty reference strip or mirrors. I've never seen one that dirty but it's mentioned in the service manual. I have seen lost documents in the ADF unit cause backside optical alarms, but the front unit is protected from this by glass. If the lamp doesn't light, it can be a defective lamp or cable. In all these cases, you'll need a tech onsite. Disassembly of this model is difficult & reassembly is even worse. I wouldn't recommend it even if you're mechanically inclined.
Posted on Apr 08, 2009
yammy_alf
  • 666 Answers
SOURCE: backend optical error
check on feeder side assembly, there might loos contact on bulb or on bulb controller.
Posted on Jun 07, 2010
I receive an error F4 on my scanner when I turn it on. Will not scan. Any suggestions?
Hello Larry,
The F-4 error in Fujitsu fi-6240 scanner indicates Background change over unit malfunction alarm. It may be caused by the malfunction of background motor or optical unit or damaged optical cable.
You are recommended to contact the nearest Fujitsu scanner service center for rectification.
Thanks & regards,
ScanCare
Sep 25, 2013 | Fujitsu Fi-6240 Flatbed Scanner
Backend optical error
Jun 07, 2010 | Fujitsu fi-4340C Flatbed Scanner
Fujitsu FL-4750 Optical alarm (g)
This is usually a main board fault. re solder the legs of the 2 sets of the 3 transistors on the underside of the mainboard. they are the small ones. or replace the main board. let me know how you get on
Sep 14, 2009 | Fujitsu fi-4750C Flatbed Scanner
Fujitsu fi5650 hispeed duplex scanner
Check the following items:
1 Does the document satisfy the paper specification?
2 Clean the Feed rollers and the Plastic rollers and see if the error is resolved.
3 Replace the Pick roller and the Brake roller and see if the error is resolved. Check the consumable counter in the software operation panel or in the built-in Maintenance mode. When the counter exceeds the values, replace the Pick roller or the Brake roller.
4 Are the cables between the Control PCA and the Feed motor damaged? Are the connectors connected properly?
5 Is the Optical unit ADF installed correctly?
6 Is the Belt ADF installed correctly?
7 Is the Belt ADF damaged?
8 Replace Feed motor and see if the error is resolved.
9 Replace the Optical unit ADF and see if the error is resolved.
10 Replace the ADF unit and see if the error is resolved.
Pls acknowledge of the issue is resolved.
Regards,
Tech Satya
Aug 06, 2009 | Fujitsu fi-5650C
Front side optical Alarm
Contaminates on the optical path to the front camera.strip and clean.If its still a problem , check and replace front lamp.(the one in the flatbed area)
Jul 04, 2009 | Fujitsu M4097D IPC Path-Through Flatbed...
E3 error on fujitsu fi-5530C
You have a rear optical error.
check the lamp is lighting.... open the ADF and push in the cover open switch. if one if off, replace it or the lamp inverter
swap the cameras over to be sure the fault follows to an E2
replace the optical unit
good luck
Jan 09, 2009 | Fujitsu fi-5530C Scanner
Fujitsu fi 5120c e2 error
See page 25 of:
http://193.128.183.41/support/scanner/manuals/fi-5120c_gs_p3pc-1312-01en.pdf
"E2" is "Trouble with the optical system (front)."
That doesn't solve your problem, but at least it narrow downs the focus of your problem.
Dec 05, 2008 | Fujitsu ScanPartner fi-5120C Pass-Through...
Fujitsu 4860c scanner error
Apr 24, 2008 | Fujitsu FI-4860C2
Fujitsu M4097D error about Front side optical alarm
HI,
I work on the fujitsu scanner hotline. This message means that the optical unit needs to be replaced, there will be very little you can do to fix this problem.
Apr 06, 2008 | Fujitsu M4097D IPC Path-Through Flatbed...
1.st step remove the software from the computer and then reboot. 2nd step install the software. 3rd step try a scan. Still in trouble? 4th step new scanner These items are not repairable.
Jul 01, 2007 | Fujitsu fi-4750C Flatbed Scanner
Open Questions:
Fujitsu Scanner Fujitsu Pad Assembly for ScanSnap S1500 + S1500M Scanners PA03586-0002 Logo
97 people viewed this question
Top Fujitsu Office Equipment & Supplies Experts
Electro Med Services...
Are you a Fujitsu Office Equipment and Supply Expert? Answer questions, earn points and help others