Question about HP 8640b Signal Generator Knob Set
Have a problem here. Generator, which is in good shape and cal, suddenly started to drop out RF output after warm up of about 10 minutes. The level meter DOES NOT drop out when this happens. After some initial success with tighening the output micro connector from the 003 option module which was way loose, it happen again. So, having a spare set of modules, I replaced the 003 module. Again, it worked for a while and then quit - same symptoms. Now I just was trying things and found that I can get the output to come back by just slightly rocking the band switch. Then I tried rocking the switch slightly to the side and leaving it there. Sure enough, the output stays on. I have a switch contact problem somewhere, but don't know where. I know that the band switch predominantly switches the ECL logic on the divider section, but am puzzeled as to why the output stays up on the meter (which is detected after the hybrid output module and right before the attenuator) indicating that output is there after the hybrid. Is there a output muting contact somewhere on the band switch that is killing the output during switching (and intermittantly when ?dirty?) ?? I do not see it on the block diagram. Thanks for your help on this.........Puzzled in Texas......Bill
  • wcarns832 Aug 02, 2010
    Some good general advice here, but obviously not familiar with the HP8640B. This is a very complex instrument. Replacing all the switches/contacts and checking all the not applicable things suggested would be a nightmare and way unproductive. If you are not familiar with the 8640B, then you can not help me. I am a retired EE, MS in EE and Physics and a very capable repair engineer. I need to know where this is.
  • wcarns832 Aug 03, 2010
    Helpful
  • 905 Answers
Hi Bill,
My experience of HP gear is limited, so bear with me.
A previous generator that I used, original 8940, did have the "muting" function as you described. Per Agilent, it is a design feature to prevent an unloaded output (mismatch) unless specifically configured. I never was able to get that deep into mine, as it was complete with maintenance agreement (gotta love off lease equipment!), and was serviced by Agilent Tech. Services. I also seem to remember an output disable function if the external reference (10MHz) was lost or drifted too far.
Out of curiosity, is your generator also configured with the option for Automated Testing (ie assembly line)? I forget what the option was for it, but it was available with rackmount kit and moved the outputs to the rear (2 sets of outputs total) and each set was electrically selectable / enabled via GPIB.
Posted on Aug 03, 2010
    Thanks for the reply and the info. No, no automatic test function on the 8640B.......Don't I wish.
    I will continue to pursue the mute thing. Something is shutting off the output after the output hybrid and detector diode for leveling feedback. Not much after that.
    I am going to explore more this AM.
    Bill
  • HP Master
  • 12,061 Answers
If it were me i would simply replace ALL switches/contacts on spec, I would imagine you have some sort of build up on the switch contacts, this may be acting as a resistor or maybe even a crude Diode... or as you suggest perhaps a contact that's not quite solid, it's making contact OK, but at a higher resistance than it should be, IE: It's acting like a series resistor? thus altering the current characteristics?
Also check your power supply for AC/RF ripple, check ALL electrolytic caps nearby. Also all RF filtering, and coupling/decoupling components.
Also check out ALL decoupling caps on the IC's you may find you have a leaky one?
It seems to me that something is damaging that 003 module, this is evidenced by the fact, that it was fine, then deteriorated,. I would imagine you have something that is slightly damaging, or causing to go out of spec, ECL is very susceptible to earth eddies, so ensure ALL earth connections are solid too. Spray ALL connections with a good contact cleaner and ensure ALL contacts are good and tight.
As for the meter not indicating as it should, there may be some sort of back feed to this meter, through a faulty decoupling decoupling component, this then would explain why it is doing what it does.
Hope this may help.
Posted on Aug 02, 2010
Trying to setup a cobra 29ltd chrome. Put on channel 20, switch set on cal, keyed up the mic, and when I turn the swr cal knob the needle won't move. Magnetic ant with acc outlet for power Any ideas
First, make sure that the CAL/SWR/RF switch is set to CAL. If you still can't move the needle, check the RF output. If the needle moves to the far right, you at least have output power, and something is wrong with the SWR circuitry. Does the needle move with incoming signals?? If the needle doesn't move, then you either have a damaged meter. If the needle moves on rec eived signals but not when you transmit and the switch is set to RF, then you most likely have a damaged transmitter output stage.
Oct 01, 2016 | Cobra 29 LTD CB Radio
Icom ic 7600 rf issue
it is just because your cable is loose, try it after tightning it up
Jun 05, 2012 | Icom IC-7600 HF/6 Transceiver
I have a Cobra CB Radio and was wondering if I have it on SRF or SWR or CAL
S/RF/SWR/CAL SWITCH. When in the S/RF position, the meter swings
proportionally to the strength of the received signal. When transmitting,
the meter indicates relative RF output power.
When in the "CAL" position, the SWR meter can be calibrated by adjust-
ing the "SWR CAL" control to the "CAL" mark on the meter face.
When in the "SWR" position, the standing wave ratio is measured.
You will want to operate on S/RF position.
Jul 14, 2011 | Cobra 29 LTD CB Radio
7 hp brigs engine on my generator runs two fast producing 127 volts. how do I slow it down?
Hi and welcome to FixYa. I am Kelly. Since you did not mention the make or model number of your generator I will respond in generic terms. If your generator is at 127 volts with NO load it is adjust at the very top end of the manufacturers limits. There is a 10% tolerance so 120 VAC + 12 is 127 VAC. Under an electrical load above 500 watts the generator should drop the output voltage slightly so your generator would still be within factory limits. Personally on any generator as a repair person; I set the output voltage at 126 VAC no load so that when the user puts the generator to work the output voltage is good. If your generator is running at 127 VAC under load then you can adjust the governor screw on the governor plate CCW about 1/2 turn to drop the voltage. Without a generator model number or engine number I can not tell you exactly where the governor plate is on your engine / generator. The screw will have a lock nut that you must loosen 1/4 turn then turn the adjustment screw CCW to drop the voltage. Just make sure to re-tighten the lock nut when the adjustments are complete. Mind you by decreasing the output voltage you also drop the output freqency at the same time you decrease the output voltage. It is better to have 125 -127 VAC under a no load condition as the more electrical load you add the more voltage drop you will notice. It is best to adjust the generator to 122.5 -123 VAC under a medium load. (50% of output capacity) A bread toaster is 1400 - 1600 watts and is a good load for a 3000 watt generator. Also a good hand held hair dryer will use around 1600 watts. If you use this type of load and the voltage output is 122.5 to 123 VAC I would not adjust anything. I have been working on generators since 1970 so this is nothing new for me. I hope that you understand everything I posted. If not, feel free to ask as many questions as you wish in any future response.
Thanks for choosing FixYa,
Kelly
Jan 08, 2011 | Craftsman Electrical Supplies
No heat at idle
radiator coolant level and overflow tank low fill overflow tank to max remove radiator cap top off coolant level, start car and let it warm up for 5 or 10 minutes, if fluid level drops top off radiator with coolant again check for heat output if OK replace cap DONE else have someone increase engine RPMs (around 3500rpm) top off coolant level (there was air in system) replace radiator cap
Dec 06, 2010 | 2005 Nissan Altima
Onan generator dies after 2-3 hours of running , wiil not restart rigt away wait 10 minutes starts no problem then runs for 15 minutes shuts down and keeps reapeating on 15 off 10
This sounds like overheating. These small generators don't have overheating protection, but if a low oil pressure or low oil level is present it may be the reason for stopping.
Check the oil level in the starting conditions and again when it stops. It may become under the "stop" level and after some time the precipitated oil has reached the "start" level. The same may be with oil viscosity and oil pressure.
Try to add some oil or change the oil with bigger viscosity.
Jul 07, 2010 | Watts Onan Portable Generator - 4000 ,...
John Q president cb
If by calibration, you are referring to the check if your antenna is tuned, below information may be of help.
When in the "CAL" position, the SWR meter can be calibrated by adjusting
SWR METER. Measures standing wave ratio of the antenna system.
Used to properly adjust the length of the antenna and to monitor the
quality of the coaxial cable and all RF electrical connections. If there is any
degradation whatsoever in any of the foregoing, due to humidity, salt
spray, vibration or corrosion, the SWR meter reading will rise, thereby
indicating that a problem exists.
To calibrate, switch to the "CAL" position, transmit by pressing the mike
switch, and adjust the SWR control to the "CAL" mark on the meter; then
switch to "SWR" position for the SWR measurement.
If however you are referring to the internal calibration or tuning of the transmit and receive sections, you would need several instruments/equipment such as RF generator (0-30Mhz with 1Khx modulation), Spectrum Analyzer, Modulation Scope, Dummy Loaded Wattmeter, Frequency Meter and some others. These of course are not readily available to the average user.
Hope this be of initial help/idea. Pls post back how things turned up or should you need additional information.
Good luck and kind regards.
Thank you for using FixYa.
Jun 12, 2008 | Cobra 29 LTD CB Radio
I can't pick up anything...
Hello even though everything is connected properly, have you checked for standing waves on the antenna to radio connection? Also you may have to perform a sensitivity test on the unit. The problem is you need a RF generator for that. Set you RF generator to the frequency that is equal to the CBs middle frequency and modulate the CW with 1KHz at 50% modulation, tune to the middle of the CB band on the radio and set the output of the RF generator to .1uv and loosely couple the signal to the CB antenna with a loop of wire. Your CB should be picking up a 1KHz tone gradually turn the power level down and when the tone cuts off that level is cutoff level, then increase the level until you can hear the tone again when the tone cuts on that value is the sensitivity of the unit. Compare the measured value to the value in the owners documentation. if the sensitivity is low this could denot a problem in the RF/IF section of the CB.
Oct 04, 2007 | Uniden Bearcat Pro PC68LTW 40-Channels...
HP 8640b Signal Generator Knob Set Logo
117 people viewed this question
Top HP Radio Communications Experts
John Trevino
Tim Whalen
John Tripp
John Tripp
Are you a HP Radio Communication Expert? Answer questions, earn points and help others